DISCUSSION

Dalam dokumen Pub1310_web.pdf(14.71 MB) (Halaman 141-148)

ADVANCES IN OPTICAL WATER ISOTOPE RATIO MEASUREMENTS

7. DISCUSSION

Optical isotope ratio measurements have clearly matured to the point where the deuterium analysis of water has become competitive with IRMS in terms of accuracy. Here, fractionation effects during sample handling constitute the dominant source of analytical errors. Although, it appears unlikely that an optical method will be able to reach the very high accuracy (~0.03‰) possible with IRMS for δ18O determinations on natural samples, current optical instruments are coming close, and are certainly able to satisfy the needs of many users. For highly enriched samples the precision and accuracy of laser spectrometry have already been shown to exceed those of IRMS. The possibility of using laser spectrometry to measure δ17O easily may be used in relation with the mass-dependent fractionation relation [28] for independent verifi cation of δ18O measurements on the same sample [15], or to verify the applicability of this relation. Or in altogether more exotic applications, such as the determination of the 17O concentration in the heavy water used in a solar neutrino experiment [29].

A growing number of spectrometers use CRDS, or one of its derivatives.

This is a logical choice for an ultra-sensitive spectrometer for the isotope analysis of water at the extremely low water concentration encountered in the upper troposphere and lower stratosphere, but may not be so for an instrument designed to measure water at more elevated concentrations or liquid samples.

One reason is that such an instrument is almost inevitably built with just one gas cell. Sample and reference are therefore measured sequentially, increasing the likelihood of a temperature difference between the two measurements producing an apparent isotope shift (typically several ‰ per degree K). For this reason we have implemented the possibility to measure a second H16OH feature with a very different line strength temperature coeffi cient, thus providing a build-in gas thermometer by which to correct the measurements.

Of course, precision and accuracy are not the only factor to be considered when comparing laser spectrometry to IRMS. There are many applications for which other aspects, like robustness, portability, remote operation in a hostile environment, compactness, and the real-time measurement capability, are more important. Other advantages of optical methods that may be considered are the high selectivity, thus avoiding the need of chemical sample pretreatment, the non-destructive nature of the measurement, making it possible to recover the sample if needed, the absence of consumables (like an ionization source

fi lament), and, low cost. Last but not least, methods based on direct absorption are conceptually simple, which helps reduce the required scale correction and normalization with respect to IRMS.

The arrival of a number of commercial suppliers of optical spectrometers signals the acceptance of end users of the technique. The near future will surely see even more precise and accurate, as well as more sensitive, devices. Also, more attention will be paid to sample handling and calibration strategies.

ACKNOWLEDGEMENTS

We are indebted to past and present students and post-doctoral fellows who work(ed) in our laser spectrometry laboratory, and our colleagues Harro Meijer and Henk Visser. We also like to thank all our collaborators, and in particular Sigfus Johnsen, Gianluca Gagliardi, Livio Gianfrani, Daniele Romanini, Marc Chenevier, Samir Kassi, and, last but not least, Hans-Jürg Jost. Substantial fi nancial support was obtained from the Dutch Foundation for Research on Matter (FOM), the Royal Academy of Arts and Sciences (KNAW), and the University of Groningen.

This paper is dedicated to the memory of Henk Visser, who passed away much too early on June 3, 2007, after a courageous and inspirational battle with cancer. His great enthusiasm and love for science will continue to guide us.

REFERENCES

[1] KERSTEL, E.R.Th., MEIJER, H.A.J., Optical Isotope Ratio Measurements in Hydrology (Chapter 9), in: Isotopes in the Water Cycle: past, present and future of a developing science, P.K. Aggarwal, J. Gat, and K. Froehlich (Eds.), IAEA Hydrology Section, Kluwer (2005) 109–124.

[2] KERSTEL, E.R.Th., Stable isotope ratio infrared spectrometry. Handbook of Stable Isotope Analytical Techniques, P.A. de Groot (Ed.) Elsevier (2004) 759–787.

[3] EPSTEIN, S., MAYADA, T.K., Variations of O-18 content of waters from natural sources, Geochim. Cosmochim. Acta 4 (1953) 213–224.

[4] BEGLEY, I.S., SCRIMGEOUR, C.M., High-precision δ2H and δ18O measurement for water and volatile organic compounds by continuous-fl ow pyrolysis isotope ratio mass spectrometry, Anal. Chem. 69 (1997) 1530–1535.

[5] PHILLIPS, A., FOUREL, F., MORRISON, J., Oxygen-deuterium isotopic measurements using a variety of pyrolysis methods — IRMS continuous-fl ow techniques, Isotopes in Environ. Health Stud. 36 (2000) 347.

[6] BRAND, W.A., AVAK, H., SEEDORF, R., HOFMANN, D., CONRADI, Th., New methods for fully automated isotope ratio determination from hydrogen at the natural abundance level, Isotopes Environ. Health Stud. 32 (1996) 263–273.

[7] MEIJER, H.A.J., Isotope ratio analysis on water: a critical look at developments, New Approaches for Stable Isotope Ratio Measurements (Proc. Advisory Group Meeting, 1999) IAEA-TECDOC-1247, IAEA, Vienna (2001) 105–112.

[8] BERGAMASCHI, P., SCHUPP, M., HARRIS, G.W., High-precision direct measurements of 13CH4/12CH4 and 12CH3D/12CH4 ratios in atmospheric methane sources by means of a long-path tunable diode laser absorption spectrometer, Appl. Opt. 33 (1994) 7704–7716.

[9] KERSTEL, E.R.Th., TRIGT, R. van, DAM, N., REUSS, J., MEIJER, H.A.J., Simultaneous determination of the 2H/1H, 17O/16O, and 18O/16O isotope abundance ratios in water by means of laser spectrometry, Anal. Chem. 71 (1999) 5297–5303.

[10] KERSTEL, E.R.Th., GAGLIARDI, G., GIANFRANI, L., MEIJER, H.A.J., TRIGT, R. van,, RAMAKER, R., Determination of the 2H/1H, 17O/16O, and

18O/16O isotope ratios in water by means of tunable diode laser spectroscopy at 1.39 µm, Spectrochim. Acta A 58 (2002) 2389–2396.

[11] GIANFRANI, L., GAGLIARDI, G., BURGEL, M. van, KERSTEL, E.R.Th., Isotope analysis of water by means of near-infrared dual-wavelength diode laser spectroscopy, Optics Express 11 (2003) 1566–1576.

[12] TRIGT, R. van, KERSTEL, E.R.Th., VISSER, G.H., MEIJER, H.A.J., Stable isotope ratio measurements on highly enriched water samples by means of laser spectrometry, Anal. Chem. 73 (2001) 2445–2452.

[13] TRIGT, R. van, KERSTEL, E.R.Th., MEIJER, H.A.J., MCLEAN, M., VISSER, G.H., Validation of the DLW method in Japanese quail at various water fl uxes using laser and IRMS, J. Appl. Physiol. 93 (2002) 2147–2154.

[14] KERSTEL, E., PIERSMA, T., GESSAMAN, J., DEKINGA, A., MEIJER, H., VISSER, H., Assessment of the amount of body water in the Red Knot (Calidris canutus): An evaluation of the principle of isotope dilution with

2H, 17O, and 18O as measured with laser spectrometry and isotope ratio mass spectrometry, Isotopes Environm. Health Studies 42 (2006) 1–7.

[15] TRIGT, R. van, Meijer, H.A.J., SVEINBJORNSDOTTIR, A.E., JOHNSEN, S.J., KERSTEL, E.R.Th., Measuring stable isotopes of hydrogen and oxygen in ice by means of laser spectrometry: the Bølling transition in the Dye-3 (South Greenland) ice core, Ann. Glaciology 35 (2002) 125–130.

[16] SPEAKMAN, J.R., Doubly Labelled Water, Theory and Practice, Chapman &

Hall: London (1997).

[17] SPEAKMAN, J.R., The role of thechnology in the past and future development of the double labeled water method, Isotopes Environm. Health Studies 41 (2005) 335–343.

[18] KERSTEL, E.R.T., Van der Wel, L.G., MEIJER, H.A.J., First real-time measurements of the evolving 2H/1H isotope ratio during water evaporation from plant leaves, Isotopes Environm. Health Studies 41 (2005) 207–216.

[19] BERDEN, G., PEETERS, P., MEIJER, G., Cavity ring-down spectroscopy:

Experimental schemes and applications, Int. Rev. Chem. Phys. 19 (2000) 565–607.

[20] MAZURENKA, M., ORR-EWING, A.J., PEVERALL, R., RITCHIE, G.A.D., Cavity ring-down and cavity enhanced spectroscopy using diode lasers, Annu.

Rep. Prog. Chem., Sect. C, 101 (2005) 100–142.

[21] WEBSTER, C.R. HEYMSFIELD, A.,Water Isotope Ratios D/H, 18O/16O, 17O/16O in and out of Clouds Map Dehydration Pathways, Science 302 (2003) 1742–1745.

[22] KERSTEL, E.R.T., IANNONE, R.Q., CHENEVIER, M., KASSI, S., JOST, H.-J., ROMANINI, D., A water isotope (2H, 17O, and 18O) spectrometer based on optical- feedback cavity enhanced absorption for in-situ airborne applications, Appl. Phys.

B 85 (2006) 397–406.

[23] MOYER, E.J., HANISCO, T.F., KEUTSCH, F.N., SAYRES, D.M., ALLEN, N.T., St. CLAIR, J.M., WEINSTOCK, E.M., SPACKMAN, J.R., SMITH, J.B., PFISTER, L., BUI, T.P., ANDERSON, J.G., Measurements of the HDO/H218O relationship in stratospheric water vapour and implications for the cause of stratospheric isotopic enhancement, In preparation (2007).

[24] LEE, X., SARGENT, S., SMITH, R., TANNER, B., In situ measurement of the water vapour 18O/16O isotope ratio for atmospheric and ecological applications, J.

Atm. Oc. Techn. 22 (2005) 555–565.

[25] AGGARWAL, P.K. GROENING, M., GUPTA, M., OWANO, T., BAER, D., Laser spectroscopic analysis of stable isotopes in natural waters. AGU poster, abstract H51D-504 (2006).

[26] FUSCH, CH., SPRIRIG, N.,, MOELLER, H., Fourier Transform Infrared Spectroscopy Measures 1H/2H Ratios of Native Water With a Precision Comparable to that of Isotope Ratio Mass Spectrometry, Eur. J. Clin. Chem. Clin. Biochem. 31 (1993) 639–644.

[27] GRIFFITH, D., personal communication (2007).

[28] MEIJER, H.A.J., LI, W.J., The use of electrolysis for accurate δ17O and δ18O isotope measurements in water, Isotopes in Environm. Health Studies 3 (1998) 349–369.

[29] KERSTEL, E.R.Th., O-17 Analysis of Ontario Hydro Heavy Water, Customer report No. 2001-32, Centre for Isotope Research, Groningen, The Netherlands (2001).