Chapter 4: Dual-wavelength cavity ring-down spectroscopy for methane isotope ratio measurement
4.5 Discussion
Our data acquisition scheme is shown in Figure 4.3. Figure 4.3A shows a raw spectrum, a composite of two overlaid spectral regions with every other data point coming from one laser. Consequently, parsing the overlaid spectrum results in two independent spectrum, as shown in Figure 4.3B. The total wavelength coverage using our acquisition scheme is 1.54-1.67m, which covers 12CH4, 12CH3D and 13CH4 isotopes.
Figure 4.3: A. Spectrum collected from our dual-acquisition scheme. The spectrum is composed of two spectra from two independent lasers overlaid. B. The spectrum for the two regions (~1.66m and ~1.54m) after parsing the data. With this approach, we can access distant regions of the methane spectrum simultaneously for our isotope ratio measurement (limited only by the cavity mirror reflectivity curve, which was measured to have R>99.995% from 1.54-1.67m).
This acquisition scheme allows us to acquire high signal-to-noise spectra, scan over a broad wavelength range nearly simultaneously, and automate the process to reduce user intervention. A critical factor for precise spectroscopic isotope ratio measurements is the ability to choose rotational lines with similar ground state energies and measure them simultaneously. To this end, we chose the pairs of lines to measure 12CH3D/12CH4 (red),
13CH4/12CH4 (blue) isotope ratios shown in Table 4.1. The samples were enriched for
13CH4 and 12CH3D (composition denoted in Figure 4.4 caption) due to dynamic range limitations. The positions, intensity and ground state energies are obtained from the HITRAN 201221 database. These choices were made based on optimal ground state energies (to minimize temperature dependence), and minimal spectral interference in our spectral region. The last column shows the spectrometer’s detection sensitivity in parts
per-billion (ppb) (101.325 kPa and 298 K) for a SNR = 1:1 for the selected methane isotopes transitions in natural abundance (0.0156% D, 1.1% 13C content).
Isotopologue Position (cm-1) HITRAN Line Intensity (cm/molec.)
E" (cm-1)
Number Density (molec./cm3)
Projected Sensitivity
(ppb)*
12CH4 6006.06590 6.060x10-24 293.154 9.05x1014 12
12CH3D 6457.03268 8.149x10-27 46.554 1.89x1014 16,373
12CH4 6004.86265 3.245x10-22 10.482 4.78x1013 0.6
13CH4 6008.46523 5.793x10-24 10.482 2.04x1013 25 Table 4.1: Selected methane isotopologues for D and 13C measurements. The position, line intensity and ground state energies (E”) are found in the HITRAN 201221 database. The number densities for the measured transitions were calculated from the fitted spectral areas and the HITRAN line intensities. *The last column shows the spectrometer’s projected sensitivity at 101.325 kPa (1 atm) in parts-per-billion for these transitions at natural abundance (0.0156% 12CH3D, 1.1% 13CH4 content).
Single-line fits are shown in Figures 4.4A and 4.4B. The fitted SNR using the Galatry profile for these spectra ranged from 2000:1-5000:1. Spectral interferences limit the fitted precision to below that shown for the isolated 12CH4, 13CH4 spectral fits in Figure 4.2. For all cases, spectral fits were constrained with fixed Doppler widths calculated from measured temperature and pressure-broadening parameters (self and N2) reported in HITRAN 2012.21
Figure 4.4: A. The spectra and VP/GP fits for the selected D/H isotope ratio pair are presented. The samples composed of an enriched composition of 12CH3D:12CH4 (10:1) mixture with a total pressure of 100 mTorr (0.0133 kPa) B. The spectra and VP/GP fits for the selected 13C/12C isotope ratio pair are presented.
The 13CH4 transition is centered at 6004.8626 cm-1 and the 12CH4 transition is at 6008.4652 cm-1.The sample composed of N2:13CH4:12CH4 (100:1:1) mixture with a total pressure of 100 mTorr (0.0133 kPa).
The pressures ensure lines were well isolated to facilitate spectral fitting. On average, the spectra fitted SNR, defined as the peak absorption divided by the RMS of the residuals, ranged from 2000-4000:1 and 3000-5000:1 for the D/H and 13C/12C ratios, respectively.
The isotope ratio precision of our method was tested by 14 consecutive acquisitions of D/H and 13C/12C isotope ratios. Each ratio pair (13C/12C or D/H) acquisitions requires 1500-1800 seconds for 300 ring-down averages, 200 MHz frequency steps, for a total acquisition time of >7 hours for the ensemble. In both cases, an isotope ratio is determined from a ratio of the integrated areas for each isotopologue, with values of the Doppler width, Lorentzian width, and narrowing parameter computed and fixed for all of the fits. During the acquisition period of one isotope ratio pair, the average standard deviation of the cavity temperature was measured to be 30 mK, which was used to propagate temperature uncertainties of the measured line intensities. Figure 4.5 shows the modified Allan deviation (defined below) for a representative measured ensemble expressed in terms of the notation in units of per mil (‰).
sample 3
standard
(R 1) 10
X R
(eq. 1)
Here X is the minor isotopologue of interest, and Rsample is the sample isotope ratio of the rare isotope to the abundant, and Rstandard is the similar isotope ratio of a conventional standard (Vienna Pee Dee Belemnite for 13C and Vienna Standard Mean Ocean Water for deuterium).
To quantify our measurement of precision, we define the quantity of the modified Allan deviation, Allan, in terms of the isotope ratio acquisition number n:
2 1
( ) 1
2(N 1)
Allan i i
i
n n n
(eq. 2)Here, N bin size and n iis the average value of the measurements in bin i. The results of the Allan deviation with respect to the bin size N is shown in Figure 4.5.
Figure 4.5: A. The Allan deviation for the ensemble of values measured for D/H isotope ratio. B. The Allan deviation for 13C/12C isotope ratios. For the D/H case, the Allan deviation for the acquired is minimized for 7 acquisitions (~3 hours); for 13C/12C, 14 acquisitions (~ 7 hours).
In the case of the D precision, the Allan deviation reaches a minimum value of
~0.26‰ with a slope of =-0.3(1). For the 13C, the minimum Allan deviation is ~0.11‰
with a slope of =-0.3(2). While white noise limited measurements follow the power-law
relationship with a slope =-0.5, our measurement precisions for both D and 13C show
<-0.5 indicating that our spectrometer suffers from other noise sources that introduce
long-term drifts. This is especially for true for D, whose fitted uncertainty lies outside
=-0.5. The most common source of long-term drift in measuring spectroscopic line
intensity is due to temperature. Bergamaschi et al.120 gives the uncertainty in the value for isotope X, X(in units of ‰), arising from the uncertainty in the temperature, , as,
δX 2
"
1000
T
E kT
(eq. 3)
where E” is the difference in the ground state energies between the two measured transitions, k is the Boltzmann constant, and T is the gas temperature.. Accordingly,
D, is ~ 4 ‰/K while 13C‰/K, since E”=0 for the 13CH4 and 12CH4
transitions. Therefore, it is expected that temperature drifts will limit the averaging noise floor of D, but not that of 13C. We attribute this increase in the Allan deviation after 8- 9 ratio acquisitions (3-4 hours) largely to these temperature effects. Moreover, it is important to note that a D ~ 0.26‰ uncertainty corresponds to a T ~ 65mK, which is reasonable compared to the measured cavity temperature drifts. When the uncertainty of the temperature measurement is added in quadrature with the spectroscopic uncertainty, the total measurement precisions are: D ~ 0.286‰ and 13C ~ 0.11‰. From these studies, we observed that the dual-wavelength spectrometer displays long-term stability, especially over periods of greater than 7 hours of averaging for 13C that do not display large temperature dependencies.